GC baseline problems - Agilent CP 7420 column
Hello, I have a Agilient CP 7420 column connected to the GCMS-QP2020NX also equipped with a FID detector and the column is connected to the FID. The column and the chromatograph are new, but the column was purchased from 3 years ago but was not unpacked
In terms of FAME separation, I am very pleased with the column. However, I have some problems with an unstable line at higher temperatures which generally make it very difficult and sometimes impossible to quantify the analytes.
Above 175C, irregularities begin which, with the appropriate enlargement, look like 'saw blades'.
Basically, all procedures related to the chromatograph were performed (column conditioning for many hours at different temperatures, cleaning of the injector and FID, the split filter was replaced, the silica gel in the hydrogen generator was dried, the septa and liner was replaced, the column was cut several times. and reinstallation, filters are installed on both the carrier gas and the detector gas lines).
The manufacturer's service after performing all the above-mentioned steps found that these instabilities came from the column. My question is whether it is possible to deal with the fluctuations attached in the picture or Is it just the column's fault? Or maybe the injections of various solvents, such as methanol or acetone, can help. so far, n-hexane has been used
I would be grateful for any help and suggestions
https://drive.google.com/file/d/1Y3kY9U ... sp=sharing
https://drive.google.com/file/d/1A9XaNz ... sp=sharing
https://drive.google.com/file/d/1tyxZdb ... sp=sharing
https://drive.google.com/file/d/12XyOpn ... PzbS0ydXhn
/view?usp=sharing
Assuming that the column is a Cyano phase, I have seen something similar happen once with a BPX70 column. Changing the carrier gas from hydrogen to helium solved the problem.
595Comp, thank you for your answer. Based on your experience, why changing the carrier gas can help. Could the higher flow velocity in the case of hydrogen make a difference?
just noticed
the FAME analyzes are performed using Helium 1 ml/min linear velocity
the attached chromatograms were made using helium as the carrier gas
BPX70 columns come with a note that says "To fully utilize the high thermal stability of the BPX70 GC column, Trajan Scientific and Medical recommends the use of helium as the carrier gas when operating the column for extended periods
above 220/230°C." I don't think it is flow related.