high temperature reverse phase column??
9 posts Page 1 of 1
hi
could anyone suggest me reverse phase columns (C8 & C18) which can work at a range of 80oC to 95oC..???
thanks a lot
I think there is a zirchrom column out there which is made out there that goes up to 200 C.
You can use Zorbax column from Agilent. The StableBond phase C18 can be used up to 90 degrees, C8 and C3 are 80 degrees.
Hamiltons polymer based PRP-1 column (reversed phased) has been used successfully up to 225°C.
A review article about high temperature HPLC was published in LC-GC in 2004:
Stationary Phases for High-Temperature LC Separations
Yu Yang and Donald R. Lynch, Jr.
Department of Chemistry, East
C a rolina University, Greenville, NC
27858, e-mail yangy@mail.ecu.edu
LCGC LC COLUMN TECHNOLOGY SUPPLEMENT JUNE 2004, pp34-38
Robert Haefele
Dear Rick1112:
ZirChrom has a variety of chemistries that are stable throughout the pH range and up to temperatures of 200C. More information is available on our website (www.zirchrom.com) or if you would like a specific recommendation please feel free to contact me directly kellyjohnson@zirchrom.com.
Good Luck!
Kelly Johnson
Technical Support & Sales Specialist
ZirChrom Separations, Inc.
We just finished an intensive study towards the stability of stationary phases at high temperature. This will be published soon in the Journal of separation sciences.
Our findings are that the Waters Xbridge is by far the most robust silica based phase column. Columns based on inorganic supports like Zirconia are also stable, but their efficiency is very poor. I would not recommend these columns.
hi
just wondering..what happens to stationary phase and bonded phase at high temperature????
This totally depend on the conditions of the mobile phase. Under neutral and basic conditions, the silica dissolves. Under acid HT conditions the silica dissolves and the ligands are stripped off. In addition the content of organic modifier also plays a role.
High temperature LC and phase stability are both relative definitions. When is a column stable??
Porous graphitic carbon columns can also go higher in temperature and pH 1-14 without problem.
9 posts Page 1 of 1
In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)