PE Turbomatrix 350 ATD

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Does anyone have experience with this instrument. We have one and I have been getting leak fail error, not from the cold trap but the tubes. For a while it would fail the first time and pass the second, now its not passing at all. It is not the tubes. I have tested the flow of the splits through the tube. Inlet split will not keep a constant value. Outlet and desorb splits are just fine. The service representative said it sounds like a blocked line and that he will come out a replace the carbon traps and resolve the issue for $4000! This seems a little much for our company. If anyone has any ideas on what I can do, either schematics of the instrument or knowledge where the carbon traps are, or how to fix the blocked line without it costing so much I would appreciate the help.

Hi there.

Did you check the O-ring? Both the small and the big O-ring that are seated in between the inlet and outlet. Hope you understands which one that I am referring to.

When we encountered a leak, what we usually did with the ATD system is to conduct a normal PM and usually this will solve the issue.


All the best !!!

Yeah, I've checked and replace the two in the upper seal as well as the ferrules. I very carefully tightened things back. I'm pretty certain about the blocked line. I just don't know how to fix it.

I see, you dont have the schematic or the service manual with you?

Actually, I have an older version of the ATD ; let me check the manual first and get back to you on how to service the blocked line - hope the setup is similiar with the newer 350.

BTW, we sometimes encountered leak issue again even though we replace new seals and o-rings. Perhaps, you can replace with other rings and seals again; just to double check?

I will get back to you soon, hopefully. Cheers...
4 posts Page 1 of 1

Who is online

In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Semrush [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry