Stability indicating HPLC method for Minocycline
Can anyone please refer me to a stability indicating HPLC method for testing minocycline and its degradation products? Any publication?
Or any advice?
Thanks!
Use an acidic mobile phase (0.1% formic acid or 0.1% phosphoric acid) with methanol. Detection at about 260 or 360 nm.
A standard reversed phase column may work, but I had better separation with a Zorbax SB-AQ column. Try about 50% methanol. I was able to see the main peak and two main impurities.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
the hplc method for determiation of it has been published in one of chinese journal . the abstract of the method as follows:
mobile phase:water(20mmol/l ammonium acetate pH=6.5):ACN:MeOH=830:185:5
colum type:C18
wavelength:280nm
looking for laboratory job in USA
Here is method used by one of our customers for Tetracycline stability method:
http://www.sielc.com/compound_285.html
Your compound is from the same class of compounds and can be analyzed in a similar fshion.
Vlad
This is excerpt from Waters application note book.
http://www.waters.com/webassets/cms/lib ... try168.pdf
Best regards,