Your point 2: sadly, new does not necessarily mean clean. Try vials and caps from a different manufacturer. I agree that the autosampler is probably contiminated, but then the question becomes where did that contamination come from - unless you eliminate the source it will just come back again. Is there a reason that you are nut flushing the whole system with pure methanol, which you have already demostrated by injections of solvent blanks has the ability to mobilise the contaminant.
Peter