Autotune issues Agilent 6470 LC-QQQ
Hoping someone else has come across this issue. We only run in negative mode on an Agilent 6470-QQQ. Our instruments do not hold a calibration. When going to autotune for negative mode, there is a failure for ion 302. This happens on both of our systems. This has happened with multiple lots of tune solution. We can sometimes get it to go away by cleaning the capillary but this is now not solving the issue after completely replacing the capillary. Any feedback is much appreciated.
Recently I had similar issue with 6410 QqQ, but with ions with larger masses. The trouble was in nebulizer needle, after heavy cleaning and depth adjustment, autotune was without errors. The primary err was: "Adjust mass axis failed: delta is outside the tolerance level". The second problem was in MS2 freq, that one was ruled by service, just software adjustment via telnet. I'm new to the forum. Hope this helps.
Make sure you are using a fresh and high-quality tune solution appropriate for negative mode tuning. Ensure that the solution is not expired and is stored correctly according to the manufacturer's recommendations. melon playground
You mentioned that cleaning the capillary sometimes helps. Since you've replaced the capillary already, ensure that the new capillary is installed correctly, and there are no leaks or blockages. pizza tower
Dealing with autotune issues on the Agilent 6470 LC-QQQ can be frustrating, especially when it seems like a persistent problem. It's great that cleaning the capillary has helped before, but replacing it without a resolution is concerning.