Setting retention time windows for EPA 300.0

Discussions about IC and related topics

2 posts Page 1 of 1
I am new and running EPA 300.0 on an ICS-6000 and I'm trying to make sure the processing methods are following the DEP guidelines. After calibration, the current retention times are set in the component table tab. But as far as the retention time windows, how is that set? The method states "3 times the standard deviation of the retention time" so am I adjusting this percentage each time as well to that result? Or is that formula for something else?
Forgive my inexperience but I can't seem to find an answer in the software or elsewhere.
Please include a mention of the data system you're using. ICS 6000 speaks to several CDS platforms...

My suspicion is that you make 3 or 4 injections, calculate a mean value and standard deviation for each peak of interest and set your retention window to be approximately RT±3sd, and then you can forget it.

Depending on your CDS, there may be a convenient box to check to accomplish this in your processing method.
Thanks,
DR
Image
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 0 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 2 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry