waters 991 PDA
I am looking for software for the 991 pda.
Also data cable and manual would be helpful.
The 991 has not been supported for a number of years (like 10) so users guides could be hard to come by. If I remember correctly it uses a IEEE cable to control. As for software, you are looking at finding an old version of Millenium. You may be able to get a signal out of the analog output but I think if you want it to work as a true PDA with control you may be out of luck.
I might have a manual for that...and it should run on an old Millennium system. The cable is an IEEE-488, I think. If I recall correctly, it's not the most sensitive detector - the 996 is about an order of magnitude quieter and I'd recommend one of those highly. With that said, the 991M was a reliable machine.