Column Pressure too high
Page 2 of 2
Unfortunately i cant change column temperature on this machine
I would guess that since it happened right away, that you've crudded up the head of the column.
It probably isn't a lipid problem (your wash cycle would have taken care of that) but more likely you've got proteins and other bio-junk strongly absorbed on the column. You probably need to flush the column with detergents, try SDS or a denaturant like guanidine thiocyanate.
You should re-consider your sample preparation stage - maybe try 0.2um filtration instead of 2um. If the 5um column worked OK then maybe switching to the 3um is like asking for trouble.
Page 2 of 2
In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)