How to communicate with GC HP5890siriesII through rs232 port
Dear GC expert,
I had two GC HP-5890 siriesII (from donation) but donator don't have GPIB card and its cable and I can't fine supplier (because it out of support and I'm not in USA). Please tell me how to communicate with HP 5890 through sr232 port and more detail about it such as what cable, what computer port, software and driver, how to set dise board.
Thank you for help every comment.
Yut.
You need a board for the GC, I believe it was an Option 580 board. This installed in the GC and gave both HPIB and RS-232 communications. Agilent probably doesn't have this board anymore, try the major third party suppliers and you should be able to get this.
I can offer the following three business contacts as possible sources for the HP cables and boards that you require.
jolson@gmi-inc.com
gwarning@ttiservice.com
jkroska@netzero.net
In case the GC is equipped with HP-IB/RS-232 board and you are missing the HP-IB board for the PC and GP-IB cable , you could use our Clarity software with RS-232 control. For details, see
http://downloads.dataapex.com/documenta ... t-5890.pdf
The board for GC could still be obtained at second hand instrumentation suppliers, but is getting rare.
Ivan Vinš
www.dataapex.com
qazaqaz,
Agilent still sells all of the parts you need to convert a 5890 to HPIB configuration. I just priced it out for a customer of mine. It was around $2K. Otherwise, after market vendors galore, as noted by skunked_once. If you need additional vendors, post again and I will give you some more contacts. Finally, you can probably find the whole assembly on LabX as well. I am not sure about the card for the PC from Agilent or the RS232 software (I have always used the HPIB version) but I know those can be found in the after market.
Best regards.
In total there are 62 users online :: 1 registered, 0 hidden and 61 guests (based on users active over the past 5 minutes)
Users browsing this forum: Amazon [Bot] and 61 guests
