Chromat called " HAS "
3 posts Page 1 of 1
Sorry to bother you I am actually an engineer in the flow industry and this is my first post as I am trying to trace a particular chromat .
Does anyone know of a company or product called HAS ; it appears to be a gas process online chromatograph
I cant post the anlysis here but it appears to separate up to 60 components ..for example
i- butano
n-butano
2,2dimetil C3
2,metil C4
i-pentano
n-pentano
2,3 dimetil C4
3,3 dimetil C5
2,2,3,trime C5 etc etc
I realise butano = butane [ this is a spanish output ] but I don't recognise many of the others. All I know its a heavy hydrocarbon gas
Any help would be very much appreciated
Thank you Modbusman
looks like:
i is an abbreviation for iso-
C3 C4 and C5 represent the number of carbons in the main hydrocarbon chain
therefore:
i- butano = Iso-butane
n-butano = Butane
2,2dimetil C3 = 2,2 dimethylpropane
2,metil C4 =2-methylbutane
i-pentano = Iso-pentane
n-pentano = pentane
2,3 dimetil C4 =2,3 Dimethylbutane
3,3 dimetil C5 = 3,3 dimethylpentane
2,2,3,trime C5 etc etc = 2.2,3 trimethylpentane
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
JGK
Thank you for your input ..it's appreciated
Regards
Modbusman
Thank you Modbusman
3 posts Page 1 of 1