injector repeatabiltiy
3 posts Page 1 of 1
I would like to have a question about HPLC injector.
in our system suitability, we inject 5 injections of assay standard and then 2 injections of check standard.
We have 4 of the injections of assay standard with peak area about 3650, and the fifth one dropped to 3500. the RSD% was about 1.8%. It still passed, but ususally it is smaller than 0.5%. What could went wrong?
Thanks a lot
Not enough standard solution in the vial for the fifth injection? Yes/no/maybe?
hi
Check whether the septa of the vial is OK.Sometimes if it fragements , or goes into vial , you get some descrepancy or reason given by Danko, insufficient qty in the vial
Anirban
3 posts Page 1 of 1
In total there are 12 users online :: 2 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Users browsing this forum: Google [Bot], Semrush [Bot] and 10 guests