the orbitrap - what is its accuracy at present and does the technology permit going from maldi and the standard sample introduction? If so, what kind of time is spent crossing over?
vestel b. shirley, president
betves inc
166 norwood drive
reidsville, nc 27320

I only know the original LC-Orbitrap. Its accuracy is certainly better than 0.001u, without internal calibration and it keeps this accuracy for the whole day after a single calibration in the morning (this is in any case a very easy one-click process). It has a nice lock-mass feature using contaminants present in the solvents (phthalates etc.) or bits of silica off the column (your choice) which improves accuracy and the length of time you can keep going on one calibration.

For a better idea of how things have progressed since the early days of Orbitrap, you could try asking Thermo Fisher. They are also the best people to ask about front-ends.

From personal experience, with middle-of-the-range low MWt compounds (400-600Da) I am upset if I get an error greater than 1ppm, and I am a low-skill, very occasional user on someone else's instrument. Note that very weak peaks are measured with just as good an accuracy as intense peaks, and that accuracy does not depend on resolution (i.e. if, for speed of sampling, you use it at lower resolutions, e.g. 15,000, it is just as accurate as at 60,000).

One weakness (which may have been overcome, I don't know) is that positive-negative switching was a nightmare. It takes about 1 hour to warm up after changing over. Also, in any one scan, the mass range is limited to a factor of about 10 (i.e. 100-1000 or 200-2000 etc.). A scan of 100-2000 gives very weak signals above 1000. Again, this may have been improved.

There is at least one MALDI interface available for the OrbiTrap, there may be others. These add-on interfaces are usually AP MALDI interfaces, not the vacuum interfaces on a stand-alone MALDI.
3 posts Page 1 of 1

Who is online

In total there are 59 users online :: 0 registered, 0 hidden and 59 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 59 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry