analysis of alcohols on a HPLC
dear friends...
I have run various samples on this HPLC attached to a PDA, but this is the first time I see this:
compounds like methanol and ethanol are getting only 50-120 mAU. I have noticed most compounds get values higher than 2000 mAU. They are all standards, about 2000 mg/L.
What does this mean?
Were they poorly identified? Does this suggest a GC would analyse them better?
Method:
Mobile phase: 0.7 ml/min for 10 min.
acetonitrile 70% / water 30%.
What wavelength are you looking at? Look up the UV spectrum of an alcohol (ethanol or methanol, for example) and you will have your answer.
I know that detector technology has improved somewhat but I would still prefer to see peaks of no more than 1000 mAU.
I've seen methods with peaks at 2 -3000 mAU where calibrations are non-linear (detector overload), but with reduced injection volumes and lower Absorbances are perfectly linear.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
In total there are 34 users online :: 1 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Users browsing this forum: Amazon [Bot] and 33 guests