Hello, I am a novice in experience with Hplc, I have ecountered something that has me stumped. When I inject my sample no peaks are detected. However, if i switch the valve to the other flow, the injection is detected during the realtime. It seems like the flow lines are crossed somehow, but I never physically changed anything. Has anyone had any experience with something like this before. Thanks
My system is:
Shimadzu
2 LC 20-AD pumps
Sil- 20 HT Autosampler
Spd-20A Detector
I run isocratic flow with one pump.
What do you mean when you say "switch the valve to the other flow?"
With a 2 pump Shimadzu, you probably have 4 lines that feed the two pumps. Make sure that the proper line is selected for the proper pump (or toss all 4 lines into your mobile phase bottle...)
DR
-try to check your LC realtime in PUMP category maybe you used the other pump
-try also to check each line from the mobile phase reservoir to your pump.
- try to use gradient flow and test each individual line.
What is the backpressure before and after you've switched "the valve to the other flow"?
In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)