Internal standard for VOCs analysis with P&T GC-MS
Hello everyone,
I would like to add an internal standard to my system for the determination of VOCs (VC, cis-DCE, TCE, PCE)in water that it is not volatile but still has the properties of my compounds.
The column that I am using is a RT-QSPLOT 30mx0.32mmx10um and the absorbent of the trap is TENAX.
Lately, we had some probles of the sensitivity and the problem came from the MS, so I would like to add an internal std to control my system and I think that the best solution would be a similar compound but at the same time not so similar, not volatile.
Is anyone that could give me an advice?
If not, I think I will add a internal std in a gas phase directly to the P&T.
Thanks a lot.
Greetings,
bsxenia
bsxenia,
Bromofluorobenzene is the first thing that comes to mind. Heavy, halogenated, but volatile enough.
Since you are using MS why don't you just use something deuterated like 1,1,1-TCA?
Best regards.
Thank you very much for your answer!
It's a good idea.
Sounds to me as if it could be water quenching ionization in the ms. Using an IS will help but, it's only a band aid fix.
I use fluorobenzene as my IS , and BFB and 1,2- dichlorobenzene d-4 as surrogates.
See other posts about method 524.