what is argon purged USP ethanol? Does it mean it's purer?
4 posts Page 1 of 1
what is argon purged USP ethanol? Does it mean it's purer than regular 190 proof ethanol?
Thanks.
I don't know what "argon purged" is.
200 proof ethanol (no methanol or IPA) is available from Sigma-Aldrich or alcohol suppliers. Be aware that HPLC grade alcohol may also contain 5% methanol and/or 5% IPA, and be anhydrous or 200 proof but still not be what you want.
USP charges a zillion dollars for everything.
If taken literally I would think it means to purge the USP ethanol with argon gas, similar to a helium purge, to remove air bubbles? Now why you wouldn't just use an in-line degasser I'm not sure. Reminds me of an article by Dolan in LC Troubleshooting on Grandma's Ham.
Actually, argon purge would make bubble problems worse. I seem to recall that argon is more soluble than nitrogen. :lol:
4 posts Page 1 of 1
In total there are 5 users online :: 0 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 5 guests