accuracy solutions concentration
Page 1 of 1
the 120% and 80% of nominal concentration for accuracy study has to be exact or approximate concentration is okay?
- tom jupille
- Site Admin
I believe that's the *minimum* range over which you have to validate.
I think you should target 80 and 120%. And would not need to be exact, we aren't. Usually 90-110% (or tighter) is the spec range for us anyhow for our products.the 120% and 80% of nominal concentration for accuracy study has to be exact or approximate concentration is okay?