nilo » Tue Apr 19, 2011 12:55 pm
Hi Peter, Thanks for your fast response ^^
Hi Nilo
I still cannot see chromatograms.
A very large peak that elutes early in the chromatogram may be due to the solvent - you can check this by looking at the spectrum. In GC-MS it is normal practice to turn off the MS filament until the solvent peak has eluted, and to turn it on before the first analyte peaks elute of course. You may need to extend the time for which the filament is turned off, or it could be that the solvent peak is either eluting later than it should, or is tailing more than it should. Late elution may be due to a carrier gas flow that is too slow, or a column temperature that is too low. Tailing of a solvent peak may be becuase the inlet liner or the top of the column are dirty, or because the split vent is opening too late (or not at all) in a split/splitless injection.
>> I had set the solvent delay to 4.5 minutes and during this period, GCMS filament is off. After 4.5 minutes, the Filament on MS side will lighting. I'm using splitless mode.
About the glassware cleaning - make sure that you rinse very well with clean water after you wash with detergent, and then after you bake at 500C do not rinse with other solvents, which can only add contaminants. If you leave an open beaker of solvent to evaporate down in a fume hood it will pick up all sorts of contaminants from the air. It is better to evaporate it under a stream of clean nitrogen.
>> Great Idea!!! I had tried so many times to clean my beaker with various cleaning method but still failed to get the nice blank. I will try to perform the solvent evaporation process under stream of N2 gas. I have to rinse the beaker with Hexane, as stated in the instruction before preparing the sample (my sample need to be rinsed with hexane and then the rinsage (hexane) is collected at cleaned beaker.
And you should not trust library matches unless you have supporting evidence for the identification.
>> Just wonder why the MS library cannot identify to simple hydrocarbon compounds. For examples,
It identifys a compound as tetratriacontane which have RT at 15. Tetratriacontane have 44 carbon (if i'm not wrong). When i saw the RT, i get suprised and checked the probabilty, OMG, the probabilty above 90%. I'm using ACCU Alkane standard and feel shocked
Peter