SPE oil extraction
10 posts Page 1 of 1
hii.
i want to extract a Pharmaceuticals from olive oil samples, i don't know which kind of solvent can i use in this extraction
can you help me please?
thanks
Try David at DKLINE43@yahoo.com
hii.
thanks
C.Tony Vella Royal British Legion
WWW.HPLCworks.net
858.663 751
Arte et Marte
hii.
i try to contact David at DKLINE43@yahoo.com, Delivery to the following recipients failed.
can you send me another mail.
thanks
Opps, my friend gave you the wrong email address. It's klined43@yahoo.com. Guarantee to receive your response :lol:
Hi mmajdop,
Thought I'll answer your question here. Depending upon the solubility of the pharmaceutical you are trying to extract from olive oil. Methanol is mostly immiscible with oil, but if you prepare 10 to 50% water with methanol, you'll get near complete immiscibility with oil. On the other hand, acetonitrile is an excellent solvent to extract from oil. You just need to check the solubility and polarity of the pharmaceutical. Hope this works for you.
David
thank you all
i will trying to dissolve a pharmaceutical standards in (MeOH:H2O), adding the standards to the oil, homogenize, passing the oil sample with a standard in a SPE C18, and Elute the Compounds with acetonitrile .
I'd love to receive comments
thanks
mmajdop
Remember to think about the pH and the pKa of your compound; if your analyte are polar you could use a mix of Waters:MeOH (or MeCN) at low pH in order to extract basic compounds and high pH for acidic compound followed of hexane clean up for remotion of more polar coextractive compounds. Nevertheless; remember if you use high concentration of Orgnic solvent the recovery of your analytes could be low depending of what SPE cartideg you use. If you can extract your analytes by pH this is going to be much selective and I think the main objectvie of SPE it would br to concentrate your sample.
Look these papers.
http://www.moprn.org/Presented/Evolutio ... torage.pdf
http://www.srmr.org.in/vol1-1-2010/02.pdf
Regards.
Oscar.
I would strongly second oscarBal's suggestion if you can do it this way. Convert it to the ion, extract it with water, convert it back and extract it with solvent. If do-able, very effective.
Depending on the size of the analyte, you might consider GPC as a clean-up.
Best regards,
AICMM
hii.
thank you all,
what do you thinks is more effective C8 or C18 SPE for the samples extraction??
and whitch detector is More sensitive for the determination??
i thinks HPLC is beter than GC for the determination of this samples??
C8 will provide you with a cleaner extract for a highly hydrophobic and uncharged analyte. If only slightly hydrophobic, use a C18. C18 is less selective. Another consideration is endcapped vs unendcapped. If your analyte is highly hydropbobic use endcapped. If your analyte has polar functionalities and is weakly hydrophobic, go for unendcapped.
HPLC vs GC - largely depends on the nature of your analytes
Don Shelly
Don Shelly Consulting, LLC
don.shelly@donshellyconsulting.com
Users browsing this forum: Google [Bot] and 2 guests