Quantitation of Sodium Permagnate
I know this sounds strang, but does anyone have experiences on quantitation of Sodium permagnate using HPLC system? What kind of detector i should use for that?
It does sound strange. Why not just titrate?
The marketing team just gave a quote to a client with this suggestion, using HPLC to quantify sodium permagnate in a mixture of organic chemicals. I have no choice but to go along with the decision. Besides, the columetric method might be too labor intensive.
I have seen/read a method for ammonium ions using ion exchange column equipped with a ion detector (http://www.google.com/search?hl=en&lr=& ... ia*tobacco). I was wondering if simular approach can be done using ion exchange column for permagnate.
Since the company doesn't have any ion detector available, what other choices i have for the detector?
Suggestions?
You can do quantitation either using UV (permaganate ion) or ELSD (both sodium and pemanganate) we have a few application for simultaneous guantitation of organic and inorganic compounds.
If you would like to retain pemanganate you need anion exchange column.
For sodium you need cation-exchange column. You can retain both of them on mixed mode Primesep AB column. I will see tomorrow if we have pemanganate in the lab. May be we can do a fast method for you.
You can also check the following links:
http://hplcmethods.com/compound_081.php (sodium, zinc, chloride, etc.)
http://hplcmethods.com/compound_018.php (ions along with dextromethorphan)
You can also download the brochure about determination of ions along with organic compounds:
http://allsep.com/brochures/PrimesepIC_Rev3.pdf
And poster for "Simultaneous Separation of Organic and Inorganic Compounds in Single HPLC Run"
http://allsep.com/brochures/OrganicInorganic.pdf
Vlad
If this is permanganate it is violet and, threfore, also absorbs in the visible.
pochengjean,
I would stay away from HPLC for quantitation of permanganate. There's not much chance of it making it through a column intact. It will decompose quantitatively when retained via anion exchange (at least when the retention site is organic as in the case of Primesep)!
Producing Primatenonsep?
Users browsing this forum: Semrush [Bot] and 14 guests