Liquid chromatography capacity factor
Basic questions from students; resources for projects and reports.
5 posts Page 1 of 1
Hi
At presently I have just joined analytical method development,
I am confused how to calculate capacity factor practically.
Thank you in advance
Ashvin
1. Measure the retention time of your peak and the dead time of your column.
2. Divide retention time by dead time and subtract 1
k' = (tR/t0) - 1
Thanx ,
But how to calculate columne dead volume?
Ashvin
Ashvin Panchal
How to calculate column dead time
Ashvin Panchal
You don't calculate it, you measure it, just like you measure retention time. If you don't see a "solvent front" on your chromatogram, try injecting pure acetonitrile or methanol (I'm assuming you're using reversed phase here) and look for a small disturbance on the baseline.
5 posts Page 1 of 1
In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 15 guests