NOISE-SENSITIVITY ISSUE
We have been working on USP method for one of my products.The details are as follows.
Buffer: Dissolve 0.5 g of tetrabutylammonium hydrogen sulfate and 1.0 g of monobasic potassium phosphate in water, add 2 mL of phosphoric acid, dilute with water to 1000 mL, mix, and pass through a 0.45-µm filter.
Mobile phase—Prepare a degassed mixture of Buffer solution and methanol (18:7).
Wave length:UV 293 nm.Flow:1.2 mL/min
Column:L11-250X4.6 mm-5 micron.
We have observed very high noise using above conditions with one brand column of L11 Packing.However ,no noise was observed when using other brands of L11.Is it possible that a specific column contribute to more noise killing sensitivity of the method?Please advice.
L11 is phenyl boded phase. Yes, some brands bleed more than others, depending on the bonding chemistry and the purity of the underlying silica.