Agilent 5971 BFB Tuning Problem
I am running a 5971 MSD that has been out of service for awhile and we are trying to get running again. The entire system has been flsuhed, the source cleaned/trap changed etc. I am having trouble tuning the bfb. I can get the ion ratios where I want them (131 = 40% , 219 = 44% relative to 69) but all of the isotopes are missing (ALL of them) when I run my BFB. Any suggestions?
Is resolution OK?
Width of peak at half height?
Abs. abundance of 69amu?
What does the autotune set the multiplier voltage at ? You may need a new EM as they don't like to be exposed to air for very long.
EM voltage is around 2450 (seems high)
Peak width of the tuning peaks is .52, .53, .50 and the width at half height for the BFB peak is .07s
The abs abundance for 69 is 384640
The chromatography is a bit noisy and my peak shapes are not great either. The BFB peak shows ~1500 abundance while the noise level is around 400
You need new EM.
Are you shooting enough BFB? EPA 524 allows for 25 ng of BFB on column. How fresh is your BFB solution?
I also think you need a new EM.
working on getting a new EM now. Will report back once its installed, thanks guys.
In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Users browsing this forum: Semrush [Bot] and 27 guests