I can see from literatures that ACN and MeOH are the most common for metabolomics study, but no one mentioned the difference and perference.
Does anyone has some comparison articles for this purpose?
Thanks
I don't have an article, but for what it's worth:
Acetonitrile: lower viscosity so you can run it faster and the peaks tend to be better defined; slightly more eluting in RP chromatography; more expensive (and there was the great ACN shortage a while back).
MeOH: opposite to above, and a few things ionise in MeOH that don't ionise in ACN

Other differences: ACN turns to HCN inside you and poisons you; MeOH turns to formic acid inside you and poisons you, so both are fairly nasty for humans.
Thanks a lot for the helpful information.
3 posts Page 1 of 1

Who is online

In total there are 31 users online :: 1 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Semrush [Bot] and 30 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry