Gas bubbles at 40min
My system seeing the gas bubbles at 40min of 60 min gradient elution.
Flow rate: 0.5 ml/min
A: NaOAc buffer Ph6 B; MeOH
Shimadzu 10AVP system
Degass by He, always see bubbles at 40min no matter how long the mobile phases degassed.
Please supply some hints.
Acetone
First of all, are you sure they really *are* bubbles (as opposed to junk eluting from the column)?
Assuming they are bubbles:
Make sure that you degas continuously (not just "one shot"), because once you stop the He, your solvents begin to re-equilibrate with the atmosphere.
Keep the runs of teflon tubing between the reservoirs and the pump(s) as short as possible. Teflon is somewhat permeable to air (that's why I like the in-line membrane degassers that sit immediately upstream of the pump!).
Yes, buy a degasser to sit on top the pump. Problem solved.
It's also a good idea to mix your A solvent to the starting mixture of your gradient, and re-adjusting the gradient programme to keep the slope the same.
