Hi at all, in our lab we have Thermo TSQ ultra :cry: Instrument is used for numerous method. Some of them use HFBA for ion pairing (amine in ESI+).
Now we have serious problem with negative ion mode: background of 213,259,427,169 m/z is 10e7 and sensitivity suffer.
I'm sure it's coming from hfba sticked somewhere in system. Anyone has a good cleaning method to overcome this problem?
I've already done some maint at my system like source cleaning, capillary change, etc.
Any help will be very appreciated.
Thank you
daniele
Do you have the option of replacing the whole LC?
All standard disclaimers apply. My posts are my opinions only and do not necessarily reflect the policies of my employer.
Only for test or definitely? In first case i can attach another pump, or another hardware without problem...we have some unused equipment..from jurassic park :cyclops:
daniele
I've heard HFBA hangs up in the degasser of the LC--so cleaning the MS isn't enough.