Base sample solution and acid mobile phase
Hi there,
I am new for HPLC. I have a question here. Currently I have an method which use ammonium hydroxide (0.1%) as sample preparation solvent but use 0.1%TFA water and 0.1%TFA in ACN as mobile phase. Column is Atlantis T3.
I wonder if there is any negative impact since one is base and another is acid.
Thank you in advance.
As long as the injection volumes are small and your peaks are well-retained (i.e., away from t0) it shouldn't be an issue. If it is, it will probably manifest itself as baseline instability and/or peak shape problems.