Solvent for latex
Does anyone have any idea what kind of solvent to use on latex for GPC work? Also for Polyethylene, the recommended solvent is TCB, but seems prohibited to purchase. Any alternative solvent to use?
for polyethylene I would try hot toluene to dissolve it. I used that for additive analysis for milk bottles. It takes a while.
MEK might work.
Chromatographer1
MEK might work.
Chromatographer1
Pardon me. What is MEK? Thanks
toluene works well for most inhibited latex samples. I've seen THF work for some but toluene is better. WARM the samples for 4 hours or more if you can. Use a laser pointer to shine through the solution in a clear vial to see if you have gel floating around -- it usually sparkles or looks hazy through the focused beam. Often seen in old or uninhibited fresh latex samples.
For PE, forget toluene or MEK. You need TCB or maybe ODCB at 120C or higher to dissolve these. Hot toluene will extract the low MW fraction and possibly additives but will not dissolve the polymer.
If you can't get TCB in your area or country, you can forget the sample unless alternate techniques like thermal analysis or pyrolysis GC could tell you what you want to know. Call the suppliers below, see what they say.
Waters and Polymer Labs make GPC systems capable of operation at these temperatures, controlled from injector thru detectors.
LCguy
Thats very informative. Thanks alot!