Any recomandation which liner to use for pesticide analytics
5 posts Page 1 of 1
Hello everybody.
We are starting with pesticide analytics by GC-MS (Shimadzu)
and i was wondering if anyone could recomand liners for this purpose.
I am thinking about SKY liners from Restek
Thanks for your help
Kind Regards
Mr. Brown
Sky, there is no other answer.
If you tell us the instrumental conditions you want to use, we can think about the type.
Davide Balbo from Italy
We will use a OPTIC-4 Injection Unit for large volume injections (100 µL)
Therefore we will go for PTV injection.
We will use a Zebron MultiResidue-1 (phenomenex) column
Kind Regards
Mr. Brown
I use a ZB-5MS for a pesticide screen containing 100 analytes. This column produces consistent, reproducible results with very little column bleed. I recommend getting the one with the 5 meter Guardian column attached to prolong column life. Sky liners are great, I use an APEX liner which can accomodate large volume injections (up to ~250 uL I believe)
Mr. Brown.
Thank you for choosing OPTIC-4.
If you need any assistance, please don't hesitate to contact us.
www.atasgl.com
(i.horsting@atasgl_com) change _ to .
5 posts Page 1 of 1
In total there are 11 users online :: 2 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Google [Bot] and 9 guests