precision LC-MS/MS
19 posts Page 2 of 2
http://www.flickr.com/photos/56860579@N02/5247320768/
as the chart in the link shows that Waters ZQ2000 signal/peak area was 1) drifting and 2) had some abnormal variation when the same solution was injected multiple times. One vial was only injected once.
this was a gradient from 75%B to 100%B in 2.5 min. A: 0.1% acetic acid, B: ACN(no acid). 0.5mL/min, Water BEH 50x2.1 1.7um column, 35 C. Diluent: ACN. Analyte is a zwitterion.
I hope some one can tell me if this is normal.If not what I can do to improve it.
Thanks in advance
Excel
for system suitability testing on our LC-MS-MS (Waters 2790-ABI/4000) we used to use a <5% acceptance criteria on peak area ratio (API/IS).
For an autosampler precision qualification test once (on an LC-MS-MS system), I put a UV detector between the LC and mass spec, injected caffeine, and monitored both responses. By UV, RSD ranged between 0.1 and 0.2%. By MS, RSD was 2 to 3.3%. Convinced me to use UV detectors to qualify autosamplers....which is not at all the same as setting system suitability parameters on an LC-MS-MS method.
All standard disclaimers apply. My posts are my opinions only and do not necessarily reflect the policies of my employer.
You also have to ask...
How much precision does your autosampler give you????
Alp