Re: Establishing LLOQ and Reporting Range
Posted: Wed Sep 22, 2021 2:24 pm
I just checked some old data and fluoride works great with a quadratic fit. That's the most non-linear anion from what I see. Unfortunately we can't use it at the lab because the method states you must use linear. So the new calibration I am trying to get implemented has fluoride from 0.1 ppm to 4.0 ppm. Weighted 1/X so our reporting limit will be more likely to pass. 1/X squared worked great for the reporting limit but at 4 ppm recovery is only 94%. That's good enough but I like things within 5%.