Actually it was F8 that brought up the data processing window. F9 was autointegrate. Is there a place to put in your dilution factor?
Yes that did it. thanks. I am using 5.1.12.258 on win 10.
"Properties" and "Calculator" are actually very powerful things.
As the probe dilution factor can be used these 4 per-layer properties.
Additionally you can add custom factors named "probe_nom1" ... "probe_nom16" and "probe_denom1" ... "probe_denom16"
External standard help page in PDF https://lar.inpnet.net/support/uc/help_page_extstd.pdf
English help file (missed in distributive)
https://lar.inpnet.net/support/uc/unichrom.eng.chm
Just for fun -
This is blood alcohol headspace analysis bundled for toxicological expertise.
"Properties" - contain template for sample attribute filling (LIMS-ready export).
Some of the properties have drop-down selector which can be static - list of constants from INI-file or dynamic - i.e. database query parameterized with other properties value (select doctor name depending on selected hospital name).
All the names are fake.
"Report" - minimalistic form ready to print
"Calculator" - average of two analysis
"Calibration" - not shown - relative response ethanol/propanol vs relative conc