I can not comment on how to improved the yield of your procedure as I do not have the details - all the steps and reagents.
I am using 3% Sulfuric acid in Isopropanol vol/vol heated up at 100C for 2 hours. The matrix is standard triglyceride cocktail.
Cheers
From the information you give, I can make reccomendations of a few things to try:
You can try changing the length of time for derivitization. Be sure the isopropanol is as dry as you can get it. (The side reaction is hydrolysis of the esters - and if there is an appreciable quantitiy of water in the isopropanol, you will generate some free acid) Check the procedure for making the sulfuric acid in isopropanol.
Why use IPA for this reaction? The steric properties of IPA might not give you good results.
I would use dry methanol instead - as the derivatized products are more stable.
Also, I assume you are exracting into another solvent - like chloroform or hexane. Make sure after extraction to put the sample through anhydrous sodium sulfate to dry it. It helps to keep the samples stable.
I use IPA since Methyl esters are more volatile and I have experienced lower recovery.(Hexane). also I am mostly interested on Butryric acid which in a DB-225 column comes off early with solvent tail. I use GC-IRMS(have to keep the He back-flush open for 200S to avoid solvent to the reactor).
I am interested to know why IPA esters are less stable?