HP 5890 Series II GC oven temp at 822C
Posted: Thu Oct 24, 2019 6:49 pm
by Pie
We are using a old 5890 series II GC, and it is reading an oven temperature of 822C regardless of what the temperature is in the oven. I'm guessing it is a bad thermocouple, but I can't find any information on it. I traced it back to the main board and it plugs into the upper right corner of the main board.
Does anyone know where I can get a thermocouple, or can i just use a type K and splice it in?
Thoughts?
Pie.
Re: HP 5890 Series II GC oven temp at 822C
Posted: Thu Oct 24, 2019 7:40 pm
by antonk
As mentioned other posters - that is the right sensor:
https://www.restek.com/catalog/view/5106
This is for injector/detector
https://www.restek.com/catalog/view/4021
Posted: Thu Oct 24, 2019 10:57 pm
by andrew ryder
Isn't that the injector block heater and sensor?
https://www.restek.com/catalog/view/5106 might be what you're looking for
Posted: Fri Oct 25, 2019 12:08 am
by Consumer Products Guy
I've replaced those all those sensors; pretty straightforward.
Posted: Fri Oct 25, 2019 5:34 pm
by James_Ball
The second link is the correct one for the oven.
The trick to removing the old sensor from the plug at the main board is to use the tip of an autosampler syringe to push in the little clip that holds the square gold connector in the plastic plug, then they will slide right out. The new one should slide in and click into place, just make sure the little clip is facing outward to the hole in the side of the plug, once in deep enough it will lock into place.
The tough part may be getting the inside of the oven apart, the sensor is attached to the back panel of the oven, the four screws that hold it in can be almost welded to the fittings if the oven was used at very high temperatures.