Ethanol analysis with poor reproducibility in headspace
Posted: Wed Jan 08, 2020 1:17 pm
Hello, I am analyzing ethanol samples in aqueous matrix. However, I am not having good reproducibility. RSD% is close to 20% for ethanol. I am using dichloromethane as internal standard and this compound has 2% RSD. I believe I performed over 100 injections. Replicates are never reproductive.
I have already performed factor analysis of various parameters, but found nothing that is statistically significant. I tested the same method on other equipment, the result was not satisfactory. The peaks are well resolved, the problem is that the variation in ethanol area is so "random".
I don't know what it can be anymore. I don't know much about GC, I'm still learning a lot. Therefore, I count on your understanding and help.
Follow the method used:
Method: Headspace
Injector Temperature: 120°C
Syringe Temperature: 100°C
FID: 120ºC, air/hydrogen: 400:40
Incubate time: 20 min
Incubation temperature: 90°C
Column: SH-RTX-624 - low/mid polarity - 30,0m;0,32mm;1,8um
Sample volume: 500 uL
Split: 5:1
Linear velocity: 35 mL/min
Carrier gas: He
Sample prepare:
Stock solution ethanol: 2,5 mL ethanol P.A to volumetric baloon 25mL with ultrapure water;
Stock solution Dichloromethane: 150 uL to volumetric baloon 50mL with ultrapure water;
Intermediate Stock Solution Dichloromethane: 2mL stock solution
dichloromethane to volumetric baloon 50mL with ultrapure water.
Replicates 100ppm v/v (5): 50 uL solution stock ethanol and 2mL intermediate stock solution dichloromethane to volumetric baloon 50mL
We have already changed several of these parameters, just do not change the column and matrix.
I have the impression that it is something related to the organic composition of ethanol. Maybe adsorption interaction with some component, though, don't understand much about it yet. We didn't change the internal default either. Maybe if we change the internal standard for an alcohol, maybe the sample and the internal standard can suffer the same interference ...
Method: Headspace
Split: 5:1
Carrier gas: He
Sample prepare: