Cleaning FID
Posted: Fri Feb 07, 2020 7:10 pm
by Kaliane
Hi
Can I remove the column from the oven and heat the FID?
I want to heat the FID to clean its up.
What safety precautions do I do?
Re: Cleaning FID
Posted: Fri Feb 07, 2020 9:32 pm
by James_Ball
If you install a blank ferrule ( one with no hole in the center) you can run the FID just as if a column is installed. You may need to increase the makeup gas flow to compensate for the missing column flow if it doesn't ignite easily.
Re: Cleaning FID
Posted: Fri Feb 07, 2020 11:40 pm
by Consumer Products Guy
I'd plumb a length of stainless steel tubing and keep some carrier gas flowing ! Notice the bold !
Re: Cleaning FID
Posted: Sat Feb 08, 2020 10:34 am
by dblux_
Hi
FID is not an ECD to termally clean it.
Just dismantle your FID and clean parts with solvents (you may need a brush in case of more difficult dirts).
Re: Cleaning FID
Posted: Sat Feb 08, 2020 2:10 pm
by rb6banjo
dblux's approach is the best. The FID is easy to take apart and truly clean with solvents and sonic baths. I'd go that route if I'm you.