5977A Mass Spec keeps losing connection to acquire data.
We have a 7890B GC and a 5977A mass spec using MassHunter B.07.06.2704 and it keeps losing connection during the second run or just random.
After this occurs we cannot get any data. However, if we reboot everything works for a couple of runs than randomly stops anytime during the run.
We have swapped NIC cards around and changed out the Switch already.
Check the event viewer logs for Windows. I'm assuming its Windows 10 OS? We had FID's that did this all the time because parts of the OS were triggering and crashing the connection to the instrument.
Also, is it networked into a company network? So is the switch connected to the outside world? If so, ensure that all the IP addresses are known by the DHCP server. Your IT department can do this. Other computers may be assigned the same IP as your equipment, which will cause it to crash out. The fix is to run the outside network to a separate card and keep the instrument IP's something simple like 10.0.0.1 etc.. like the old days.
Third thing, have you upgraded the OS or the software? Was the appropriate modifications to the firmware made if you did?
Simon Wicks
Specialist Chemist
Eurofins Chemtest
