use of DCM
Can we use Dichloromethane in supelcosil LC-18-DB column?
pngeethaa
sure, why not? why would you?
Alex
thanks alex,
one of our client wants to use this column with DCM and with a gradient program of,
Time (minutes) DCM, Methanol
0 10 90
20 10 90
55 50 50
60 50 50
60.1 10 90
70 10 90
pngeethaa
Be careful if you have PEEK parts in your hplc system. I believe PEEK may have compatibility problems with DCM, but I might be wrong.
I would ask the client to explain the method to me, and also I would ask him if he ever ran this method on more than one column.
Imh has right. Mobile phases with % DCM higher than 5% can cause incompatibility issues with PEEK tubings or columns.
If it is an Agilent system that you are using, the supply different piston seals for normal phase work.
GCguy
In total there are 10 users online :: 3 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], DuckDuckGo [Bot], Semrush [Bot] and 7 guests
