Thanks to all, with 30m column i used 2.1 ml/min as flow.
For increase my sensitivity i want to change my column with 20 m and 0.3 film, change drawoutplate and lens insulator from 3 mm to 6 or9 mm.
I will say to you if the noise and background will decrease.
Nicolò
If the column is 0.25mm ID and 30m length and oven temperature is 40C then the inlet pressure with Hydrogen is 6.75psi, If the column is 0.25mm ID and 20m length and oven temperature is 40C then the inlet pressure is only 2.81psi. You will have to be careful not to back flash solvent when injecting with such low inlet pressures, also the vacuum system would have trouble pumping 2.1ml/min hydrogen so you will lose sensitivity because of higher analyzer pressure.
If you use a column that is 0.18mm ID and 20m length and oven temperature 40C then at 2.1ml/min the inlet pressure is 19psi. But you can lower the flow to 1.2ml/min and still have an inlet pressure of 10.9psi but much lower analyzer vacuum for better sensitivity. A 0.25mm ID 20m length column will not go below 1.69 ml/min with hydrogen without requiring less than 1psi inlet pressure at 40C.
If you increase the analyzer pressure you lose sensitivity, so better to go with lower flows with hydrogen if possible, but that also requires lower inlet pressures or smaller diameter columns.