PAH analysis GC-MSMS
2 posts Page 1 of 1
Dear Colleagues,
PAH analysis
GC-MS, PTV-injector (glas gooseneck liner with "frit"), He-cariergas
Problem components: indeno(1,2,3-cd)pyrene, dibenzo(a,h)anthracen and benzo(g,h,i)perylene
We have a problem with contamination in our GC-system, I think in the inlet. And have tried all sorts of things to eliminate the problem without any succes. (other methods, cleaning fase off 350 °C (PTV)). And still will the problem components appear in the solvents afterwards.
Concentration of which the problems manifest significant is of 300 p.p.b (ng/ml). Split/splitless injection with a splitless time of 3 minutes.
An other problem is that we have problems with the detection of acenaphthylene (MS-MS).
Thanks in advance for your help
"problems with the detection of acenaphthylene"
The problem is solved.
We work in SIM now.
The other problem is under control.
2 posts Page 1 of 1
In total there are 2 users online :: 0 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 2 guests