FID response difference between two Agilent 7890A GC!
Hello, everyone! Currently , we do valiations work on residual solvent method by Agilent 7890A GC system . There are two same instrument in our lab. Yesterday , we did the intermdate precision test using another 7890A ,but today morning , we found the object peaks had higher response in the second 7890A system . Compare these dates , we found the peaks height and area are about three ~four time of the first instrument! Note , the concentration of the standard solutions are same , the column are the same one , the condition are same , only the instrument is different! i am confused by these data, what's the reason?
Anyone had this experience ? or any advice for me?
thank a lot!
bily
Split ratios could be different, air/hydrogen/auxiliary gas to detector could be different. Column protrosion through ferrule into inlet could be different. Syringe sizes could be different even if softmare setting matches. Good hunting.
Also range and attenuation differences. If you are collecting with Millennium/Empower, there may also be a difference in the instrument method used, i.e. the scaling factor.
Hi
Adding sealing of the headspace vials as a possible issue.
Hi Bily!
I already transferred methods from a 6890 to a 7890 both with FID and had very similar area results. Is the autosampler also identical? Sample introduction in inlet could make a difference. Is the column distance in inlet and detector the same length in both instruments? You really have to check all the method parameters.
Good luck!
At_om
Firstly , i should say "thanks" to everyone who provided the helpful advises for me ! this problem i think must be somthing wrong in the second instrument . Even if the everything seem good in the second 7890A , but check the past chromatogram and data, the second instrument performace was normal . SO , I think the second instrument went wrong. Now, we ready to do this test in another 6890GC ( because the second 7890A is not availble now), i believe the results will be acceptable !
thanks you all again !
bily
In total there are 63 users online :: 0 registered, 0 hidden and 63 guests (based on users active over the past 5 minutes)