Broad range of active fractions
I'm purifying a natural ligand. I expect it to be ~35-55 amino acids.
Following a preliminary bulk C4 solid-phase extraction step, the first HPLC step is a C4 10 mmID column. The gradient is 27-50% CH3CN over 60 minutes (0.1% TFA). Ambient temperature. I loaded 15 mg extract (Vydac catalog claims the column can handle 50 mg). The active fraction range is Gaussian, though has some tailing. It spans t=14 min to t=40 minutes. In other words, it spans over 1/3 the chromatogram.
At first I would suspect something may be wrong with the column, however, a separate ligand we are screening for elutes AFTER the broad peaks and spans only 3-4 fractions (minutes) according to our assay.
What is the possibility that, in spite of the broad active fraction range for the first, there is somehow only one active species?
Just an idea: Maybe you have (or would have) many peaks/compounds, but inadequate selectivity thus no actual separation is causing the described “mountainâ€