PPM Limit Specification
Hi.. i have a stupid prolbem! :)
Why Standard Solution is 5 ppm?
Limit spec is 200 ppm!
https://ibb.co/f2wBGPZ
because it doesn't rely on concentrations but mass of chlorides.
mass of sample in test solution tube: 10 g/100 ml *2.5 ml = 250 mg (of sample)
mass of chloride in standard solution tube: 10 ml * 5 mg [Cl]-/L = 50 µg of Cl-
50 µg/250 mg = 0.200 µg/mg = 200 ppm
okkk thanks.. but.... why 50/250? or why a fraction and not an equivalence?
Don't know exactly what you mean.
50 µg of [Cl]- give a certain turbidity.
if the same turbidity is observed in your test solution, then you would have 50 µg [Cl]- in 250 mg of your sample, which is an amount of 200 ppm.
The test states that the turbidity is less than that of 50 µg of [Cl]-, so your amount is < 200 ppm
The test should also work if you take 1 ml of a 50 ppm standard and add 14 ml of water, or 5 ml of a 10 ppm plus 10 ml of water, and so on.
Goal is to have 50 µg of chlorine-ions in 15 ml of solution before adding the nitric acid and silver nitrate.
In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 33 guests
