FTIR
Hi everyone,
i know its not the right place to put this topic as its related to LC but still i know someone somewhere will help me out.We have a FTIR pretige 21 , a brand new system cecently with pike miracle. The scanning range is from 750tp4000. but when i go through the USP, for qualitative analysis, we need to scan fom 650. So is it ok if i scan my sample from 750 to 3800? If not, then how can i increase the range ? i already contacted the vendor but their reply is pending. Please anybody can help me out.
with regard
Donal
The IRPrestige-21 in its standard configuration (ceramic light source/KBr beam splitter/DLATGS detector) has a spectral range of 7800 to 350 cm-1. Your biggest limitation is the type of material used for the ATR plate. The lower limits of various plates used in the Miracle ATR are shown below.
ZnSe 520cm-1
AMTIR 630cm-1
Germanium 575cm-1
Si/ZnSe 550cm-1
Diamond/ZnSe 525cm-1
I've mostly used the Pike Miracle ATR with a diamond/ZnSe plate and have had no problem scanning down to 525cm-1. It does get a bit noisy close to this limit because there's not much energy getting through.
In interests of full disclosure, I'm a service engineer for Shimadzu.
Hope this helps.
In interests of full disclosure: then I'm mad at you "by proxy". Shimadzu telephone and E-mail help literally cut me off, refused to return calls or E-mails concerning communications issue with Shimadzu GC autosampler on an Agilent 5890 GC, the autosampler wouldn't communicate with ChemStation after a few years. So Shimadzu is on my sh*t list.
I'm sorry to hear about that. Not very fair being mad at me personally though!
I'm a vendetta kind of guy !!!
Like Homer Simpson said, upon hearing there was a two-day waiting period to buy a gun: "but I'm mad NOW !!!".
In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 24 guests
