Benzene in acetone
Dear all,
I have to quantify benzene impurity in pure acetone; specification limit is "not more than 2 ppm". Can I reach this low level with a GC equiped with FID and Split/splitless injector ?
Any suggestion about possible method will be apreciated.
GT
Dear GT,
Benzene is a good analyte for GC, specially with a FID. You can use the split mode without problems.
You can use this following method:
Column: DB-624 [Crosslinked 6% Polycyanopropylphenylsiloxane and
94% Polydimethylsiloxane (30m x 0,32mm x 1um)]
Carrier gás: He at 35cm/sec (~2.1mL/min)
Oven: 140oC by 20 min
Injector: Split (5:1)
Detector: FID at 240oC
==>> Rt of Benzene is 6.5min
Best regards,
Carlos Teixeira
Users browsing this forum: Google [Bot] and 8 guests