Cleaning Lipid Contamination with Organic Solvents (HPLC)
I am currently using a Agilent Technologies 1200 Series HPLC with an RPC column, which was previously used for analyzing organic compounds containing high concentrations of lipids. I am getting a lot of noise in my analysis even after the standard wash procedures. The solvents used in my HPLC are (Solution A) 5%ACN/95%H20 and (Solution B) 95%ACN/5%H20.
I am currently replacing Solution B with 95% dichloromethane but this does not seem effective so far.
Any recommendations?
If you replace the column with capillary tubing: is the noise better, worse, or the same? That should tell you a lot.
A bit more detail would be useful:
is your method isocratic or gradient?
what detector? if UV, what wavelength?
is the noise "random", or is there a pattern to it?
do you get the same noise pattern when you inject a blank?
if random, is the noise higher or lower under no-flow conditions?
In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Users browsing this forum: Google [Bot] and 26 guests
