Autosampler error
I keep getting this error (5890 GC, 7673 injector):
HP3365
Failed to put vial down at front injector
Retry Cancel
Of course I click on "Retry" and the sequence continues on - but this has happened about 3 times this morning already.
There was a large storm here last night which affected some of the network (data system, email), so maybe a power surge/burp?
Any suggestions on what I should check to prevent this from happening?
John
Shut everything down, including GC and control box. Shut down computer. Start all. We just had a similar issue. I even had last week the 7673 drop a vial, had to look for it. It sure couldn't then put the vial down !!!
I went ahead and shut everything down yesterday (restarted computer, GC and control box) and started another batch of samples.
Came in this morning and I got the same error after the first sample ran(out of 30!), which apparently happened about 15 minutes after I left the building yesterday, d'oh!
:x
John
Hi John
Is this a real error or a false one ? in other words is the arm really failing to release the vial, or is it releasing the vial but its little silicon brain thinks that it isn't ?
In the first case I'd check for mechanical faults. Given the probable power spike the second one might need a software reload, or some new electronics :cry:
I actually never saw what happened when the error occurred.
I restarted everything again, and when I started up the instrument, the software dialog box read "Injector not ready."
The cable at the back of the controller was loose. Instrument is running fine now.
:roll:
With past experiences, I'd rather this be the problem rather reloading the software (do you know where yours is? haha) or installing new circuit boards
John!
Users browsing this forum: Bing [Bot] and 5 guests
