Error C0 Corona CAD Detector
I have an older version of Corona CAD (ESA/Thermo) which gives me C0 (C zero) error when self testing, which I am unable to find in the manual. Anybody knows what this is?
Thank you.
Vlad Orlovsky
HELIX Chromatography
www.helixchrom.com
https://www.researchgate.net/profile/Is ... ual_EN.pdf
Here's a link from researchgate.net to the manual
I did not see this specific error in the manual, unless it is in the section C of troubleshooting. It states for error 0 - Thermal Organizer not detected
Vlad Orlovsky
www.helixchrom.com
"C" is a hexadecimal error code that stands for "12". Adding "0" results in Error "120" which is Error 100 (IonTrap OFF --> 24V Power Supply issue) and Error 20 (Pressure Transducer 2 --> Interface Board issue).
A this device is rather old, you'll face the problem to get replacement parts and someone from THermo to service it.
Unfortunately it's better to look for the Corona Veo model.
Greetings
LCWGE
Greetings
LCWGE
Thank you very much for your input.
Vlad Orlovsky
www.helixchrom.com
