Finnigan Magnum: Major water and air leak?
I just acquired a used Finnigan Magnum ion trap MS that is giving me headaches.
The unit was in use until 2007 when it was left open to atmosphere.
The auto tune in 2007 set the multiplier voltage to 2700 V.
At that voltage I am seeing enormous peaks at masses 16-20 (one broad peak) and sharp, distinct peaks for 28 and 32 (Nitrogen and Oxygen), almost none at 40 (Argon).
I found and fixed one major air leak at the manifold flange but that didn't improve the situation.
Spraying the whole unit with argon didn't show any other leaks. The transfer line is sealed with a piece of fused silica tubing closed with a septum.
I'm currently using an RV8 forepump which gives a pressure of 5*10^-2 Pas without He being connected. The turbo pump reaches 101 percent speed after 5 min. and gets around 35°C warm after running for two hours with the manifold at 130°C.
Does the MS need Helium for a correct scan or for pumping away all the possible residue 'goo' in the trap? (The unit looked perfectly clean when changing the flange).
What pump-down times are normal for a Magnum that was left open to atmosphere for some time?
And is the multiplier voltage maybe too high?
Is it dangerous for the multipliers or the filaments to operate them with a possible leak?
Thanks in advance!
The multiplier voltage is very high. I consider a mutilplier with a voltage above 2000 at the end of it's useful life. To determine if you have a true air and water leak I would tune the instrument, and look at the ratio of 18 to 69 and 28 to 69 to see what the relative intensities. If there is a notebook of saved tune reports you can compare to see how the current ratios compare to past history.
Finally some news :)
There was no leak, just a much too high multiplier voltage. After pumping away the water from the trap the auto tune set the multiplier voltage to 1750 (quite reasonable), AGC target to 22000, filament emission current to 15 ya.
So the spectrometer is basically working.
Thanks for your help!
HbJ, we are both located in Germany and since you are running a Magnum and I'm running a Saturn 3, we seem to have similar oldtimer problems. private messages are impossible here, so if you like you can contact me at < email removed >.
Best regards,
Stephan